Topic
-
For making of a current injection source with 2000 A rating current, can we use an ordinary current transformer with ratio 2000/1 A, and connect 1 A side to one ordinary low current injection source ( 1 A) for obtaining of 2000 A in other CT side?
For making of a current injection source with 2000 A rating current, can we use an ordinary current transformer with ratio 2000/1 A, and connect 1 A side to one ordinary low current injection source ( 1 A) for obtaining of 2000 A in other CT side?
Viewing 8 replies - 1 through 8 (of 8 total)
Viewing 8 replies - 1 through 8 (of 8 total)
- You must be logged in to reply to this topic.

Assume we can produce 2000 A current in short circuited paths. In this case the primary short circuit drop voltage can be calculated as follows: V= (2 * 2000) / (56 * 2000) ~ 0.01 V And for consumption burden we can write: S= 0.01 * 2000 = 20 VA < 25 VA ( CT burden ) Is it possible? Can we produce 2000 A in primary side?